posted by Mathieu Gisselbrecht at Oct 27, 2006 4:05 PM
Quote
After 2 months of beamtime, it seems that the MCPS are working. It is not so good, but at least at high bias current (round 2300 V) there is a rather homogeneous answer. It seems however, that one side is answer much less (DS C &D). The same diagnostic was made at the lab, with the dark count only. Of course, during the experiment other effect ought to be considered...Anyway, after every baking out, the current has increased quite a bit. The new table is given here:
V | 1000 | 1100 | 1200 | 1300 | 1400 | 1500 | 1600 | 1700 | 1800 | 1900 | 2000 | 2100 | 2200 | 2300 | 2400 | 2500 | 2600 |
---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|
micro A | 32 | 35 | 38 | 41 | 44 | 48 | 51 | 54 | 57.5 | 60.5 | 63.5 | 67 | 70 | 73 | 77 | 80 | 84 |
posted by Mathieu Gisselbrecht at Sep 6, 2006 6:45 PM
Quote
This morning the dark count was less then 2 Hz @ 2100V, I increased from the voltage up to 2250V and the dark count stayed around 5Hz during all the day. I decreased the threshold of the MCPs to 83 mV (check ?) The background pressure is around 4.10^-7mbar.
During the day, I prepare the acquisition computer. I had to re-install it since the OS got frozen very quickly ???
I did some one acquisition on the dark count at 2250V, checked a little the code. Everything was fine ! Meanwhile Olivier fixed 3 dead channels from the DS16 cards, and I performed again some acquisition on the dark count and finally reached 83 % of localization. We did a quick check with the gage the count rate with a gain of 2200V. There is some work to do for time calibration now…
posted by Mathieu Gisselbrecht at Sep 5, 2006 10:14 AM
Quote
The day of yesterday has been very good until, I decided to switched off the system. High dark count rate occured just then. Today, the background pressure is arround 4.10-e7 mbar (in 3 days), slightly better then yesterday. The polarisation scheme is as followed:
With respect to yesterday, I can notice that there is 1-2 micro A more above 500 V. Courant leak ?… A first pulse with a repetition rate of 1Hz is detected with 1600 V and the anode at 1300V. The pulse width is about 10 ns and the amplitude around 15-35 mV and is independant of the gage status (on / off). It repetition rate seems to be less with the votlage on the anode around 1100 V. A second pulse with a width <5 ns and and amplitude of 5-10 mV is present for a voltage of 1900 V. At 2000V the darkcount is around 30 Hz while yesterday it around a few Hz !The system had a strange comportement during the morning and finally the reference channel (D16) died during lunch ! In the afternoon, something had changed, the pictures where much better and we could start to increase the DS16 threshold. The detector was still noisy around 500Hz (+/-300Hz) , and we let it run the whole night at 2100V.
- front MCP : -1000 V -> -14000 V
- rear MCP : +1000 V
- anode : +1350 V
V | 100 | 200 | 300 | 400 | 500 | 600 | 700 | 800 | 900 | 1000 | 1100 | 1200 | 1300 | 1400 | 1500 | 1600 | 1700 | 1800 | 1900 | 2000 | 2100 | 2200 | 2300 |
---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|---|
micro A | 2 | 4 | 7 | 9.5 | 11.5 | 14 | 16 | 18 | 20.5 | 23 | 25.5 | 28 | 30 | 32 | 34.5 | 37 | 39 | 41.5 | 44 | 46.5 |
posted by Mathieu Gisselbrecht at Sep 5, 2006 10:08 AM
Quote
After one week of work, I started up the system. The polarzation scheme is symmetric :
- MCP front : +1000 V ->1400 V
- MCP rear: -1000V
- Anode: -1350 V
posted by Mathieu Gisselbrecht at Sep 5, 2006 9:47 AM
Quote
The diagnostic from August (the 8th) have shown that there was very poor electrical contact between the feed-through and the anode. During the week (27th of August-> 1rst of september), we replaced the gold-wires glued with sivler epoxy (Epotek 21D) with regular SnPb weldings. After three iterations (including the super-sonic bath) the system was tested with RF successfully, and set up under vacuum. The HV test showed again clear limit with the feedthrough above 1100 V:
- The front side of the MCP can stand up to 2.6 kV; micro discharge is seen around 1.3 kV at 1Hz. The origin and location is unknow.
- The rear side of the MCP can stand up to 1.1 kV; above this there is a high load of the oscilloscope
- The anode can stand up to 1.4 kV; above this there is a regular "clac-clac" noise in the experimental room
- the Cu ring where very much oxyde
- the Cu ring for the grid where at a floating potential,
- the electric holder for HV where very close to the MCPs, and potentially could induce short circuit
- the front MCP was "burned" on the side probably due to a discharge...