The set-up used in order to callibrate MCPs is given by the figure below. Monitoring the voltage and the current (see table), one can determinate the MCPs internal resistance and the actual MCP difference of potential. Table : Applied voltage vs current
V100200300400500600700800900100011001200130014001500160017001800190020002100
micro A2.54.579.511.51416.5192123.52628.5313335.53840.54345.54850

In the configuration where the arrow is set the typical voltage through the plates is 1850 V. When the gage is on the count rate is about a 5kHz. A first test of imaging shows that the pulse height is two weak (small) to trigger the DS16s. In the figure below we can see that only a few card from the rear-side anode is answering (cards: E-F & G-H). Only the cards A & B from the front-side anode are present, and they have a little amount of counts. The hit distribtion is only 2 pixels: 1x +1y with an anode @800V. Since the discriminator levels are set to the minimum it means that there is not enough gain !!!… Concerning the localization of the hits on the detector it seems, that the charged particles emitted by the ion gage are localized almost in the center of the detector : image of the 35mm diameter of the flange with lens effect due to strange electrosatic fields inside the vacuum chamber, lost of gain on the edge of the MCPs ....

posted by Mathieu Gisselbrecht at Aug 1, 2006 12:00 AM